GSI Forum
GSI Helmholtzzentrum für Schwerionenforschung

Home » Alice » Alice TRD Chamber Tests » Chamber testing manual version 3
Re: Chamber testing manual version 3 [message #1384 is a reply to message #1342] Fri, 11 February 2005 17:08 Go to previous messageGo to previous message
Chilo Garabatos is currently offline  Chilo Garabatos
Messages: 12
Registered: August 2004
Location: GSI
occasional visitor
From: lxial22.gsi.de
Dear Rainer and all,
we started preparing our applications to write the data in the format that is proposed in the chamber test manual, and therefore looked with some detail at the proposed information and format.
The information is very complete, and I would like give you some comments and suggestions aimed at making life somewhat easier:
General comment: In every site about 10**2 chambers will be tested. To make things as automatic as possible is then desirable; in particular, the information that is gathered for the data base -and for our own files- should be written by the computer, not typed by the users (tedious and dangerous).
Suggestions:
Headers: Do not write the Ar and CO2 flows separately (the composition is yet another entry), and do not write the set and actual values at the beginning and the end of the measurement (3 lines with 2 reals each) but
Write the total 'high' flow and total 'low' flow (2 lines with 1 real each). This is the information that is needed for the leak rate values, which can be calculated automatically by pressing a button in the application when the O2 plateaus (the user then only types the flows and clicks).
Leak rate data: the relevant information are the leak rates at both flows and the data themselves, so write this at least (time and O2, not the other way around, as is proposed) and leave the testers the freedom to write a zero or -1 for the fit stuff and the reference O2 levels (extra work not needed according to our experience).
Gain: write at the end of the file the data in x-yyy format:
voltage, current, rate, gain (one line per point).
In this way, the number of points doesn't matter, the data base can deal with that.
It is also easier for the tester to copy-paste the data into his/her own spreadsheet and make plots for meetings.
This strategy -the data points at the end- should be applied to all files (O2, scans, ...).
Best regards,
Chilo
 
Read Message
Read Message
Read Message
Previous Topic: Radiator dimensions
Next Topic: 55Fe spectral measurement
Goto Forum:
  


Current Time: Thu Mar 28 14:15:38 CET 2024

Total time taken to generate the page: 0.01107 seconds