Chamber testing manual [message #1022] |
Sun, 10 October 2004 15:46 |
Rainer Schicker
Messages: 9 Registered: August 2004 Location: Phys Inst Heidelberg
|
occasional visitor |
From: *physi.uni-heidelberg.de
|
|
A first draft of TRD chamber testing manual is available
at www.physi.uni-heidelberg.de/~schicker/trdtest/chamb_v1.pdf
|
|
|
Re: Chamber testing manual [message #1199 is a reply to message #1022] |
Mon, 29 November 2004 19:47 |
Harald Appelshaeuser
Messages: 39 Registered: March 2004 Location: GSI
|
continuous participant |
From: *ikf.physik.uni-frankfurt.de
|
|
Dear Rainer,
the manual looks very good! I have a few questions and comments:
1.1 is the 90Sr source really needed for anything?
2. Chamber leakrate: the maximum allowed leak rate should be specified.
3. Chamber conditioning:
- I think the conditions need to be specified: flow rate, required O2 and H2O purity. Is the full drift voltage applied?
- O2 and H2O content should enter the database.
- can you specify when the conditioning was successful? what is the maximum allowed dark current, at what voltages?
4. Gain uniformity:
- which radiation source should be used (and why)? If X-ray tube: which settings are recommended?
- Conditions: flow rate, O2 and H2O? With or without drift voltage? Important: which anode voltage (sag!)?
- Which level of non-uniformity can be tolerated (standard deviation, maximum deviation)?
5. Gain curve:
- at which flow rate, O2, H2O? Which source? Which drift voltage?
- what is the recommended DeltaV and Vmax?
- should an absolute gain measurement be performed for every chamber (how much effort is this?)? Very important, but also in general: do we insist on a specific gas mixture?
Regards, Harry
|
|
|