|
Re: leakrate measurement of 4 chamber stack [message #1016 is a reply to message #1014] |
Tue, 05 October 2004 19:10 |
Chilo Garabatos
Messages: 12 Registered: August 2004 Location: GSI
|
occasional visitor |
From: lxial22.gsi.de
|
|
Hello, Sasha, hello David,
When you evaluate the leak rate of this 4 chamber stack, you could check the consistency of this measurement with the measurements of the chambers alone plus the contribution of the interconnecting bellows; 2 of this bellows have an upper limit of 1.8 +- 0.3 ml/h, as I measured at GSI with Bernd's boxes.
About the speed of flushing, I have to check a few things at CERN to be able to answer. It will be probably difficult to beat Sasha, but we'll know after filling the small chambers.
Regards,
Chilo
|
|
|