Re: Trigger OR [message #18050 is a reply to message #18048] Mon, 23 March 2015 14:49 Go to previous messageGo to previous message
Damian Ralet
Messages: 35
Registered: July 2014
Location: Darmstadt
continuous participant
From: *gsi.de
Hi Alberto,

There is a priority encoder on the trigger. The trigger with the highest number have the priority over the lower number. For example, if you have two accepted triggers, 8 and 9, the trigger 9 will be the one triggering the data read-out, and the event will be labelled with a trigger 9.

For each trigger, you have a hit-pattern that allows you to check coincidences, and see if you had more than one trigger option.

For your second question, the number of particle can be estimated with the trigger 10, multiply by the scaling factor. It has the priority. You could also look into the particle scaler module, but I am afraid that this number is not reliable. If I remember properly, we observed a lot of variations when we were looking at the scaler estimation. (Maybe someone else could comment on this?)

Cheers,
Damian

PS:If you want to read more about the trigger generation, I would suggest to check the publication:
http://www.sciencedirect.com/science/article/pii/S0168900215003332
 